Channel 0 is sampled every 200 ms. The others are sampled every 400 ms since every other measure (200 ms) is discarded to "purge" the registers. This is because the ADE7953 has low-pass filters that need some time to settle, so when switching from one multiplexed channel to another you need to account for this.

The allocation of the sampling is something I worked quite a bit, since I did not like the idea of static allocation/division among channels. Now it is a dynamic allocation based on the absolute value and variability of the power. This means that constant low-value channels get sampled less often (in any case no more than every 15s), while active channels get their priority.